Scoring the spectrum of deposited energy in a thin silicon sensor, radiative corrections and high energy simulations #826

Answered by Jaap1957
Jaap1957 asked this question in Physics
Discussion options

You must be logged in to vote

To summarize the remakrs of Mr Townson and Mr Tessier:

To add radiative compton corrections, follow the instructions here:
https://www.reddit.com/r/EGSnrc/comments/ecuql7/radiative_compton_corrections_on/
The result is a Makefile like the attached file (remove the .txt) (this one is in egs_phd):
Makefile.txt

For simulations above 1 GeV, change the 999 entries in\HEN_HOUSE\src\get_media_inputs.mortran at lines 838 and 847 to, for example, 10100 (10 GeV).

The pegs4 tables at HEN_HOUSE\pegs4\density_corrections\... reach up to 10GeV.

Replies: 6 comments 25 replies

Comment options

You must be logged in to vote
2 replies
@ftessier
Comment options

@ftessier
Comment options

Comment options

You must be logged in to vote
12 replies
@ftessier
Comment options

@ftessier
Comment options

@Jaap1957
Comment options

@ftessier
Comment options

@Jaap1957
Comment options

Comment options

You must be logged in to vote
10 replies
@mainegra
Comment options

@ftessier
Comment options

@mainegra
Comment options

@ftessier
Comment options

@Jaap1957
Comment options

Comment options

You must be logged in to vote
1 reply
@ftessier
Comment options

Answer selected by Jaap1957
Comment options

You must be logged in to vote
0 replies
Sign up for free to join this conversation on GitHub. Already have an account? Sign in to comment
Labels
None yet
4 participants